S-Meter Calibration
Is there a way to calibrate the S-Meter such that it matches reasonably to what my yaesu 3000 ft-dx and kenwood 870 exhibit? I primarily use 40 meters and have an average of an s-4 noise floor. I notice using my 7000 mk ii the average noise floor sits around s8 to s9. This is on a fully reset database and fresh install of thetis. Is some sort of pre-amp turned on? Any way to match to the other 2 radios? THx kindly.
Re: S-Meter Calibration
Signal Generator.
I use an XG3.
https://elecraft.com/products/xg3
You're more likely to find your other two rigs are wrong...as most all rigs are notoriously bad on smeters adcuracy.
You'll also proabably find your other rigs smeters depend on the RF gain which is the wrong way to design an smeter.
Mike W9MDB
I use an XG3.
https://elecraft.com/products/xg3
You're more likely to find your other two rigs are wrong...as most all rigs are notoriously bad on smeters adcuracy.
You'll also proabably find your other rigs smeters depend on the RF gain which is the wrong way to design an smeter.
Mike W9MDB
Mike W9MDB
Re: S-Meter Calibration
Hello,
I use the Elecraft XG3 Signal Generator too, like Mike. Perfect tool! And the calibration in Thetis works excellent too.
Main difference between Icom/Kenwood/Yaesu rigs and (ANAN) SDR's is that the Japanese rigs use 3dB/S-unit until S9, and 6db/S-unit above S9. You have to take this into account when you read the S-meter of these rigs.
73, Hans PA0Q.
I use the Elecraft XG3 Signal Generator too, like Mike. Perfect tool! And the calibration in Thetis works excellent too.
Main difference between Icom/Kenwood/Yaesu rigs and (ANAN) SDR's is that the Japanese rigs use 3dB/S-unit until S9, and 6db/S-unit above S9. You have to take this into account when you read the S-meter of these rigs.
73, Hans PA0Q.
--
73,
Hans Remeeus (PA0Q/OE3JRC)
https://pa0q.nl
"Communication is about people, the rest is technology"
73,
Hans Remeeus (PA0Q/OE3JRC)
https://pa0q.nl
"Communication is about people, the rest is technology"
Re: S-Meter Calibration
Yes, you will find that the big three, Kenwood, Yaesu and Icom, all cheat very badly when it comes to S-meter accuracy and linearity. They all want to claim they have the "quietest receiver". What they all really have is the "quietest S-meter"
This is a particularly good link to review: https://vu2nsb.com/radio-systems/amateur-radio-station-ham-shack/radio-transceiver-s-meter/
As you can see from the data at that link, a true signal or, more typically, noise level of S4 is reported by these radios as an S1. Hence your noise level can be quite high but you placidly think how wonderful your receiver is because it is reporting an S1.
The only receivers that have anywhere near accurate S meter readings are those from Flex, Apache and Elecraft. It is not unusual for new owners of those makes to complain about how "noisy" their receivers are because the S-meter readings are true and correct (dare I say honest?), when in fact those are some of the best receivers on the market.

This is a particularly good link to review: https://vu2nsb.com/radio-systems/amateur-radio-station-ham-shack/radio-transceiver-s-meter/
As you can see from the data at that link, a true signal or, more typically, noise level of S4 is reported by these radios as an S1. Hence your noise level can be quite high but you placidly think how wonderful your receiver is because it is reporting an S1.
The only receivers that have anywhere near accurate S meter readings are those from Flex, Apache and Elecraft. It is not unusual for new owners of those makes to complain about how "noisy" their receivers are because the S-meter readings are true and correct (dare I say honest?), when in fact those are some of the best receivers on the market.
-
- Posts: 58
- Joined: Thu Oct 14, 2021 5:23 am
Re: S-Meter Calibration
Thx for the replies, I went to the calibration menu and changed the default -70db level reference to -80db reference at 5.000 mhz WWV and at least have a closer representation to what I'm used to. I'm not after exact perfect values, but enough to match the two other radios. Thanks !
Re: S-Meter Calibration
I'm curious - why would you want to match what is known to be inaccurate - i.e. not correct? The factory setting was I'm sure very close to correct from the get-go. It is calibrated such that S9 = -73 dbm. Note all other manufacturers - ok, the Japanese, match this as well as seen in their reviews in QST. So they ARE the same at that level - the level you changed. From there going up S9+ values they continue to be quite similar.... but below S9 it falls apart, the further it is below S9 the more inaccurate the other radios are. For example, S8 on the Anan is S7 on the other radios... S7 is S5.... S4.5 (between S4 and S5) is S0 on the other radios! So with them, you THINK you are really low noise - but the reality is it's S4.5 or -100dbm.
By changing the calibration point you only changed the single point where they agreed and now BOTH would be inaccurate, the slope is different for each and thus the inaccuracy will still be there.
What I will do is I'll tell people that their report is an "accurate S meter reading on a calibrated meter".
Also: be sure you have the RX Meter set to "signal" and not "signal avg".
Gary
K9RX
By changing the calibration point you only changed the single point where they agreed and now BOTH would be inaccurate, the slope is different for each and thus the inaccuracy will still be there.
What I will do is I'll tell people that their report is an "accurate S meter reading on a calibrated meter".
Also: be sure you have the RX Meter set to "signal" and not "signal avg".
Gary
K9RX
-
- Posts: 58
- Joined: Thu Oct 14, 2021 5:23 am
Re: S-Meter Calibration
What good is the S-meter if most of my signal contacts, even weak ones, are S9 + 10 over? This is why... I can't respond to my contacts telling each and everyone of them you're 10 over lol. I'll reset the database yet again and see if the S-meter is more reasonable upon reset.
Re: S-Meter Calibration
S meters are not supposed to be "reasonable", they are supposed to be accurate.
Resetting your database will not make it any more "reasonable", but it may make it more accurate depending on what you did to the calibration.
If you have access to an accurately calibrated signal source, you can further improve the accuracy of the S meters. Based on my own experiences, and some anecdotal ones I have noted in the past, people have typically seen around a 3dB difference after calibration. Note that this is an offset only and has nothing to do with linearity. Our S meters seems to be very linear, i.e. a 6dB (S unit) change in signal strength is reported properly at all signal levels.
However, at the end of the day, providing S meter based signal reports really isn't the best method of describing signal strength and quality. It would be better to provide everyone with a signal-to-noise (SNR) report. SNR is really what matters. If somebody is S9 and you neglect to tell them you have an S8 noise floor they may think they are perfectly readable when in fact they are not.
Resetting your database will not make it any more "reasonable", but it may make it more accurate depending on what you did to the calibration.
If you have access to an accurately calibrated signal source, you can further improve the accuracy of the S meters. Based on my own experiences, and some anecdotal ones I have noted in the past, people have typically seen around a 3dB difference after calibration. Note that this is an offset only and has nothing to do with linearity. Our S meters seems to be very linear, i.e. a 6dB (S unit) change in signal strength is reported properly at all signal levels.
However, at the end of the day, providing S meter based signal reports really isn't the best method of describing signal strength and quality. It would be better to provide everyone with a signal-to-noise (SNR) report. SNR is really what matters. If somebody is S9 and you neglect to tell them you have an S8 noise floor they may think they are perfectly readable when in fact they are not.
-
- Posts: 58
- Joined: Thu Oct 14, 2021 5:23 am
Re: S-Meter Calibration
I see the conundrum. Call it old habits, Im simply not used to having ALL signals generally above S9, especially when biased with 2 other radios.
I just reset the database and will try it out again for a day. The dbm and microvolt readouts appear within rational scope as before. Guess I'll get one of those elecraft signal gens you fellas spoke of.
I just reset the database and will try it out again for a day. The dbm and microvolt readouts appear within rational scope as before. Guess I'll get one of those elecraft signal gens you fellas spoke of.
Re: S-Meter Calibration
There's another significant difference between the "big" three and SDR's like our ANAN.
The S/dBm-meter in our SDR software is completely independent from the adjustment of any attenuation, pre-amps, RF (AGC) gain, or whatever. The readout of the meter will remain exactly the same, no matter what you adjust between the ADC's and the final stage of your SDR. Our S-meter just shows the value of the signal received from the antenna lead into the ADC, nothing more, nothing less.
Now do this with an I, K or Y rig: When the signalstrength of the other station is not S9+20dB (what he wants to hear from you) you just add 20dB gain with the pre-amps and tell him the result. Friends forever!
Same you can do with your noiselevel. When it is'nt S1 you just give 20dB attenuation.
Who is fooling who?
73, Hans PA0Q.
The S/dBm-meter in our SDR software is completely independent from the adjustment of any attenuation, pre-amps, RF (AGC) gain, or whatever. The readout of the meter will remain exactly the same, no matter what you adjust between the ADC's and the final stage of your SDR. Our S-meter just shows the value of the signal received from the antenna lead into the ADC, nothing more, nothing less.
Now do this with an I, K or Y rig: When the signalstrength of the other station is not S9+20dB (what he wants to hear from you) you just add 20dB gain with the pre-amps and tell him the result. Friends forever!
Same you can do with your noiselevel. When it is'nt S1 you just give 20dB attenuation.
Who is fooling who?


73, Hans PA0Q.
--
73,
Hans Remeeus (PA0Q/OE3JRC)
https://pa0q.nl
"Communication is about people, the rest is technology"
73,
Hans Remeeus (PA0Q/OE3JRC)
https://pa0q.nl
"Communication is about people, the rest is technology"
Re: S-Meter Calibration
Good point, Hans! I believe the Flex also references all S meter measurements to the antenna jack, i.e. takes out any attenuation or preamp gain. Do the Elecraft radios work that way as well?
It's worth noting that this is how laboratory and technical test equipment works. When you look at a power measurement on an Agilent spectrum analyzer (for example), the power is referenced to the input connector and is not affected by any internal attenuation that might be switched in.
It's worth noting that this is how laboratory and technical test equipment works. When you look at a power measurement on an Agilent spectrum analyzer (for example), the power is referenced to the input connector and is not affected by any internal attenuation that might be switched in.
Re: S-Meter Calibration
The S meter used to bug me too. It took a couple of years to get over it and I still leave it on average. The Anan agrees with my HP signal generator across a wide range but it just didn't feel right.
You can use the panadapter and peak blobs to give you the readings you want, but you'll have to convert dbm to S units on the fly.
You can use the panadapter and peak blobs to give you the readings you want, but you'll have to convert dbm to S units on the fly.
Re: S-Meter Calibration
You cannot use readings from the panadapter to determine S-units except for CW:
viewtopic.php?f=6&t=2463
My old Flex 3000, and now the ANAN, finally made me "comfortable", because RF system design is my day job. I despise the so-called metering on KenYaeCom equipment.
viewtopic.php?f=6&t=2463
My old Flex 3000, and now the ANAN, finally made me "comfortable", because RF system design is my day job. I despise the so-called metering on KenYaeCom equipment.
Re: S-Meter Calibration
w-u-2-o wrote:You cannot use readings from the panadapter to determine S-units except for CW:
viewtopic.php?f=6&t=2463
My old Flex 3000, and now the ANAN, finally made me "comfortable", because RF system design is my day job. I despise the so-called metering on KenYaeCom equipment.
Sure I can. The beautiful thing about doing something wrong is making up your own rules.

The noise floor on the panadapter seems to fall more inline with the S meter on the KenYeaCom rigs.
Re: S-Meter Calibration
AG5CK wrote:The noise floor on the panadapter seems to fall more inline with the S meter on the KenYeaCom rigs.
Under average conditions (e.g. a passband of 2900Hz, and a panadapter bin width of 2.93Hz), the displayed average noise level (DANL) on the panadapter will be 30dB less than the S-meter (in Sig Avg mode). That's 5 S-units difference. Given that many of the Japanese radios show an S1 on the meter when the noise floor is really S5, that's not far off! And a sad indictment of how the marketing department probably put pressure on the engineering department

-
- Posts: 3
- Joined: Sat Oct 12, 2024 11:05 am
Re: S-Meter Calibration
All,
I have observed all this stuff while wondering why my S meter reads S9 noise level on 160m. I calibrated my S meter and checked the tracking from well above S9 down to the noise floor. It tracks within tenths of a dB. I think the real problem is the spectrum display does not show the true noise / signal peak. When I monitor Pure Signal feedback on AM my modulation peaks are also down about 20 dB compared to the carrier level. I suspect this is due to digital filtering. The only time I see things agree is in CW at 25 Hz bandwidth. I wonder if the wrong bandwidth is applied to the spectrum display. I have a Cubic radio with DSP demodulation that shows the same S meter levels on 160. Rice boxes are bogus. Frank WA1GFZ
I have observed all this stuff while wondering why my S meter reads S9 noise level on 160m. I calibrated my S meter and checked the tracking from well above S9 down to the noise floor. It tracks within tenths of a dB. I think the real problem is the spectrum display does not show the true noise / signal peak. When I monitor Pure Signal feedback on AM my modulation peaks are also down about 20 dB compared to the carrier level. I suspect this is due to digital filtering. The only time I see things agree is in CW at 25 Hz bandwidth. I wonder if the wrong bandwidth is applied to the spectrum display. I have a Cubic radio with DSP demodulation that shows the same S meter levels on 160. Rice boxes are bogus. Frank WA1GFZ
Re: S-Meter Calibration
Frank Carcia wrote:All,
I have observed all this stuff while wondering why my S meter reads S9 noise level on 160m.
It's because the noise on 160M at your location, with your antenna and station configuration, really is S9.
I calibrated my S meter and checked the tracking from well above S9 down to the noise floor. It tracks within tenths of a dB.
And there you go! The S-meter is reading accurately, and you really do have S9 noise at the antenna port on the radio. This is not unusual on 160M in urban locations.
I think the real problem is the spectrum display does not show the true noise / signal peak.
In what way do you think the spectrum display has anything to do with S-meter performance? They are making two different measurements of noise and signal power.
When I monitor Pure Signal feedback on AM my modulation peaks are also down about 20 dB compared to the carrier level. I suspect this is due to digital filtering.
Not sure what point you are trying to make here?
The only time I see things agree is in CW at 25 Hz bandwidth.
And that is absolutely correct.
I wonder if the wrong bandwidth is applied to the spectrum display.
Not at all! It's just that there needs to be an understanding that the S-meter and the spectrum display are showing two different measurements of the same thing. Once it is understood that there are two different measurements, and the mathematical relationship between them, everything will become clear. This is all explained in the following topic:
https://community.apache-labs.com/viewtopic.php?f=6&t=2463
Any radio based on, and compatible with, Thetis, piHPSDR or SmartSDR (or PowerSDR before that), is a legitimate, calibrated, spectrum analyzer. Not just a panadapter or spectral display like on the IC7300, FT710, or similar radios. Those radios provide useful displays, but they don't show any numerical values associated with the colors or lines, hence nobody notices they "don't match with the S-meter". But Thetis, piHPSDR and SmartSDR, when paired with compatible hardware, are real-deal spectrum analyzers, and you can take the numbers and the measurements on those spectrum analyzer displays to the bank. The Elecraft K4 might also be this way. This has caused all manner of confusion for hams where this has been their first exposure to spectrum analyzer operations.
In order to help with the math, some time ago Richie, MW0LGE, and Warren, NR0V, very kindly worked together to add a number of additional measurement read-outs available on the spectral display status line. In particular these are resolution bandwidth (RBW) and noise power spectral density (NPSD). If you are not seeing those then simply click once on the status line and they will be shown.
-
- Posts: 3
- Joined: Sat Oct 12, 2024 11:05 am
Re: S-Meter Calibration
Thank You!